I feel that I have not been giving enough attention to Hope2SleepGuide. Im off for the next 3 days if anyone has any questions for me. For those of you that have left me messages I promise to answer this week.

Rock

Views: 502

Reply to This

Replies to This Discussion

Many thanks for your reply. I am using a nasal pillow mask.



RockRpsgt said:

It is ppssible that your turbinites are causing problems. Very little if any though. This would be more of an issue with your daytime breathing in my opinion. What mask are you using?

 

The machines and different softwares read leaks a little diffrently. The manufactures requirements are even more stringent than my own. As a rule of thumb a leak value that is less than 4x the pressure value is acceptable to me. If your pressure is 10cmwp than you would not want your leak any higher than 40 if possible. I do give a little more freedom with Full Face masks as they tend to be leak monsters. Sometime you have to comprimise with a little leak to get a small amount of additional comfort. More importantly to watch is what your AHI is doing at different leak values on the same pressure setting.


 
snoreyhead said:

Two questions if I may?

1) I am awaiting an operation to have by turbinates, reduced is this likely to be the cause of my hypopneas being higher than my restrictive and central sleep apneas? I suffer with one blocked nostril on 98% of nights.

2) As a mask is designed to have some leakage for the purpose of exhaling when breathing what is the leak level that is acceptable?


Using a nasal mask with nasal problems has to be tough. I would imagine that a turbinite reduction would help in your case. Have you considered a full face mask until after the surgery?
snoreyhead said:

Many thanks for your reply. I am using a nasal pillow mask.



RockRpsgt said:

It is ppssible that your turbinites are causing problems. Very little if any though. This would be more of an issue with your daytime breathing in my opinion. What mask are you using?

 

The machines and different softwares read leaks a little diffrently. The manufactures requirements are even more stringent than my own. As a rule of thumb a leak value that is less than 4x the pressure value is acceptable to me. If your pressure is 10cmwp than you would not want your leak any higher than 40 if possible. I do give a little more freedom with Full Face masks as they tend to be leak monsters. Sometime you have to comprimise with a little leak to get a small amount of additional comfort. More importantly to watch is what your AHI is doing at different leak values on the same pressure setting.


 
snoreyhead said:

Two questions if I may?

1) I am awaiting an operation to have by turbinates, reduced is this likely to be the cause of my hypopneas being higher than my restrictive and central sleep apneas? I suffer with one blocked nostril on 98% of nights.

2) As a mask is designed to have some leakage for the purpose of exhaling when breathing what is the leak level that is acceptable?

Yes I have a full Respironics mask, but find the leakage and AHI increases when I use it. I don't understand the reason behind this unless its the level of leakage that renders the mask infective.

So pleased to report (for those that were watching how she got on) that June got instant relief from her bad allergy with the bacterial filters.  She's posted her own reply on The Other Discussion

Kath Hope said:

Over on the website June.  Here's the link to the pack of 5 Filters

The extra adapter needed for the Sleep Cube/IntelliPAP is HERE


June Ann Rymer said:

Thankyou Kath!!   Where can I buy the equipment I need i.e. Bacterial Filters and the separate adapter please??

A high leak will lead to an increased in AHI. I find the Respironic FFM to be the 2nd worse of all of these types of mask. Have yoou tried any of the Resmed full masks?

snoreyhead said:

Yes I have a full Respironics mask, but find the leakage and AHI increases when I use it. I don't understand the reason behind this unless its the level of leakage that renders the mask infective.

I tried at least 10 masks until I found the 'Respironics' FitLife, which surprisingly gives me the least leaks of any full face ones.  I think it's down to finding the 'right mask for the person' regardless of who the manufacturer is, which is why I think it would be good if the sleep clinics stocked ALL the manufacturers masks - in an ideal world of course, sigh....

RockRpsgt said:

A high leak will lead to an increased in AHI. I find the Respironic FFM to be the 2nd worse of all of these types of mask. Have yoou tried any of the Resmed full masks?

snoreyhead said:

Yes I have a full Respironics mask, but find the leakage and AHI increases when I use it. I don't understand the reason behind this unless its the level of leakage that renders the mask infective.

I find when I use a full face the leak and ahi are worse unless I up the pressure a little (mine is set at 8min 18 max with a swift FX or Sleepweaver, I up it to 10 or 12 min if I'm needing to use the full face mask and the Quattro FX seals much better at the higher pressure).

We have found that a FFM requires about 2cm more pressure to do the same job of a nasal mask.

symmit said:

I find when I use a full face the leak and ahi are worse unless I up the pressure a little (mine is set at 8min 18 max with a swift FX or Sleepweaver, I up it to 10 or 12 min if I'm needing to use the full face mask and the Quattro FX seals much better at the higher pressure).

Quick question about leak rate. My machine reports about 25l per minute which I am assuming is a total flow rate as I haven't got a mask programmed in, I tried a sleepweaver élan last night and it reported 15lpm, which seems quite low. Is the lower venting rate a problem and could it result in more effort being required or actually insufficient flow to supply air and vent co2?

Pressure is set at 8 to 18 and machine is an icon auto.

The masks are made to allow for proper venting of Co2. They could not be on the market in either of our countries if they were not. What pressure are you on? Resmed machine?

symmit said:

Quick question about leak rate. My machine reports about 25l per minute which I am assuming is a total flow rate as I haven't got a mask programmed in, I tried a sleepweaver élan last night and it reported 15lpm, which seems quite low. Is the lower venting rate a problem and could it result in more effort being required or actually insufficient flow to supply air and vent co2?

Pressure is set at 8 to 18 and machine is an icon auto.

Pressure is minimum 8 max 18 on a fisher and paykel con auto, 95% pressure is 9

I forgot you had the Icon like me. Those leaks are great for either mask. I would not worry.

Reply to Discussion

RSS

Sleep Apnoea Forum

New to the Sleep Apnoea Forum? 

1. Stop by our Sleep Apnoea Welcome Center to introduce yourself to the SleepGuide community.
2. Start a New Topic of Conversation.
3. Post your photos - of yourself, your old CPAP machine, your new CPAP machine, your pet, something about you!

Interested in advertising, have a problem or need to contact us? Click the Report an Issue page.

 

Latest Activity

Jonathan replied to Iain Noble's discussion Resmed Airfit F20 Woes
"Hi.  No first hand experience, but you can see reviews here, which are mixed: https://www.cpap.com/productpage/resmed-airfit-f40-full-face-cpap-mask I notice cpap.com still appear to be selling the Quattro FX.  I have imported stuff from…"
Apr 28
Iain Noble posted a discussion

Resmed Airfit F20 Woes

Dear FolksI have been using the Resmed Quattro FX mask ever since I was diagnosed over 7 years ago.…See More
Apr 28
Iain Noble replied to Becky Murkin's discussion F20 AirFit small marking bridge of nose
"Hi All I can say is I feel your pain! I got the F20 as a replacement this year and it is very uncomfortable. I have just ordered the new Resmed F40 which is very small and light and so far it has been a lot easier to use for a mouth breather that…"
Apr 28
Sarah Martin is now a member of Hope2SleepGuide
Apr 27
jjimjam updated their profile
Apr 17
Kath Hope replied to Becky Murkin's discussion noisy humidifier normal?
"It sounds like you're getting rainout issues (condensation) so here's an article to help with this. www.hope2sleep.co.uk/water-in-the-mask-hose-known-as-rainout"
Apr 16
Becky Murkin posted a discussion

noisy humidifier normal?

Hello,I have an air sense 10 auto.When I have the humidifier and the heated tube plugged in it makes a sort of drumming noise. Like that at the gallows in films.Is this normal? It wakes me up  and I have to give up and switch off  for the night.See More
Apr 16
John posted a discussion

Dentists in Hull that know about Mewing ?

I was wondering if there are any dentists in Hull who know about John & Mike Mew, and changing jaw shape over time?I could add that I saw an ENT consultant recently and she said jaw shape, malocclusion, is genetic, which has become a debatable…See More
Apr 9
Profile IconKaren Mcleod, Fiona Watson, Teresa Hobbs and 5 more joined Hope2SleepGuide
Apr 3
Jonathan replied to John's discussion Airsense 10 -humidifier runs dry
"I did replace the tank in my SleepCube, and had to regularly take the silicone sealing gasket out and give it a goof clean, but not infrequently found it standing in a pool of water in the morning.  The machine sat on a tray because of…"
Mar 26
John posted a discussion

Nasal valve collapse

Does anyone have any experience of nasal valve collapse?Wondering if I have this but I don't know what the norm is.Been trying widgets that hold the nose open and it's like night and day.See More
Mar 24
John replied to John's discussion Airsense 10 -humidifier runs dry
"Wow it's the other way round for me.I found with the Sleepcube I'd often have a good bit of water still there in the morning. Today I woke up with a dry tank on the Resmed again and it's not on full heat or flow. I got it on 6…"
Mar 24
Becky Murkin posted a discussion

F20 AirFit small marking bridge of nose

Please help or give advice all wise people on here.I have had the Resmed F20 AirFit small mask for over a month. Seals yet…See More
Mar 19
Frances Knight replied to Frances Knight's discussion Air Sense 10 - changing pressure settings
"A very helpful reply Jonathan. I found the pressure setting for the full face mask to be too high when I changed to a nasal mask, but perhaps I should have persisted? I'll try again and I'll also discuss with the clinic whether an APAP…"
Mar 13
Jonathan replied to Frances Knight's discussion Air Sense 10 - changing pressure settings
"Ramble alert!  My history is that I spent part of each week at a different location, so bought a second machine identical to my fixed pressure NHS machine and set it up to match using the clinician menu.  This was a very long time…"
Mar 12
Frances Knight replied to Frances Knight's discussion Air Sense 10 - changing pressure settings
"Thanks Jonathan - I had seen the YouTube video. My question is really about the experience of people who have control of their pressure settings, for example those people who have bought their machines directly and not obtained them through an NHS…"
Mar 12
Jonathan replied to Frances Knight's discussion Air Sense 10 - changing pressure settings
"This is NOT a recommendation http://www.youtube.com/watch?v=VNUWhd2UgQM"
Mar 11
Frances Knight posted a discussion

Air Sense 10 - changing pressure settings

I have been successfully using my air sense 10 since OSA diagnosis, and my AHI is now under 2. I had an appointment with my clinic this week and have been told there is no need for further appointments and just to contact them if I need supplies or…See More
Mar 11
Swen Greve is now a member of Hope2SleepGuide
Mar 4
Jonathan replied to Sue Reeves's discussion A few questions..
"If you have a helpful GP, I would ask them to re-refer you to the sleep clinic."